Post Reply 
bump in molecular tumbling rate-T2 relaxation
09-04-2015, 10:34 PM
Post: #1
bump in molecular tumbling rate-T2 relaxation
Why is there a "bump" in the molecular tumbling rate v T2 relaxation positive?
i.e. if T1 in T2 means that enhanced longitudinal recovery (T1) results in loss of transverse magnetization (T2) ...then when T1 was at a minimum (larmor frequency) wouldnt that "enhanced T1 recovery" have resulted in a "dip" in T2 (greater loss of transverse mag) rather than a bump upwards?

I hope my question makes sense and someone can easily answer. Thank you!
Find all posts by this user
Quote this message in a reply
11-08-2015, 06:39 PM (This post was last modified: 11-08-2015 06:57 PM by aelster.)
Post: #2
RE: bump in molecular tumbling rate-T2 relaxation
Dear kwest123,
You are right. The original figure contained an error. The caption was correct that there is a "dip" on the T2 curve, but the figure contained a "bump", not a "dip". The T2 curve should go down, not up, at the resonance frequency. I have made the correction and replaced the original figure with the corrected version below. Thank you for pointing this out!
Allen Elster
   
Find all posts by this user
Quote this message in a reply
Post Reply 


Forum Jump:


User(s) browsing this thread: 1 Guest(s)