bump in molecular tumbling rate-T2 relaxation - Printable Version +- Unfortunately, the MRIQuestions.com forum became unmanageable due to spam attacks. I recommend an alternative resource: www.imagingQA.com (http://mriquestions.com/forum) +-- Forum: The NMR Phenomenon (/forumdisplay.php?fid=3) +--- Forum: Spin, Nuclear Precession, and Resonance (/forumdisplay.php?fid=17) +--- Thread: bump in molecular tumbling rate-T2 relaxation (/showthread.php?tid=10161) |
bump in molecular tumbling rate-T2 relaxation - kwest123 - 09-04-2015 10:34 PM Why is there a "bump" in the molecular tumbling rate v T2 relaxation positive? i.e. if T1 in T2 means that enhanced longitudinal recovery (T1) results in loss of transverse magnetization (T2) ...then when T1 was at a minimum (larmor frequency) wouldnt that "enhanced T1 recovery" have resulted in a "dip" in T2 (greater loss of transverse mag) rather than a bump upwards? I hope my question makes sense and someone can easily answer. Thank you! RE: bump in molecular tumbling rate-T2 relaxation - aelster - 11-08-2015 06:39 PM Dear kwest123, You are right. The original figure contained an error. The caption was correct that there is a "dip" on the T2 curve, but the figure contained a "bump", not a "dip". The T2 curve should go down, not up, at the resonance frequency. I have made the correction and replaced the original figure with the corrected version below. Thank you for pointing this out! Allen Elster [attachment=2] |