Post Reply 
TR Question
09-23-2017, 09:58 PM
Post: #2
RE: TR Question
(09-23-2017 10:32 AM)Jerca Wrote:  Hello All,
Quick question: Using a 1.5T, If the following tissues have the following T1 relaxation times (Tissue A=420ms, Tisssue B=920ms & Tissue C=2020ms. Can someone tell me what TR will tissue B fully recover longitudinally?

The general rule is that you need about 5 X T1 to allow a tissue to recover its full longitudinal magnetization. Thus for tissue B with T1=920 ms you would need a TR of at least 5 x 920 = 4600 ms.
Find all posts by this user
Quote this message in a reply
Post Reply 


Messages In This Thread
TR Question - Jerca - 09-23-2017, 10:32 AM
RE: TR Question - aelster - 09-23-2017 09:58 PM
RE: TR Question - Jerca - 09-24-2017, 10:18 AM

Forum Jump:


User(s) browsing this thread: 1 Guest(s)