larmor frequence
|
08-26-2015, 06:06 AM
Post: #1
|
|||
|
|||
larmor frequence
Hi everybody !
Can you explain me this: if wo = gamma.Bo and if wo = 2(pi)fo why fo = (only) gamma.Bo too ???? i don't understand why it isn't: fo = gamma.Bo/2(pi) ? Thanks for any answer. |
|||
08-30-2015, 11:06 PM
Post: #2
|
|||
|
|||
RE: larmor frequence
(08-26-2015 06:06 AM)DAVID FORTIN MENKE Wrote: Hi everybody ! The gyromagnetic ratio (gamma) can be defined in terms of either angular frequencies with units (radians/Tesla) or "regular" frequencies where it has units of MHz/Tesla. I have chosen the latter. The two differ by a factor of 2 pi. See the following link if you are still confused. http://mriquestions.com/angular-frequency-omega.html A Elster |
|||
08-31-2015, 03:01 PM
Post: #3
|
|||
|
|||
RE: larmor frequence
thanks. now it's ok. Have a nice day !
|
|||
« Next Oldest | Next Newest »
|
User(s) browsing this thread: