Difference in distance measurements between encoding directions
|
11-12-2015, 05:48 AM
Post: #2
|
|||
|
|||
RE: Difference in distance measurements between encoding directions
Dear Xen,
The most likely cause of the problem is not your sequences, but one or more miscalibrated gradients. Gradients commonly drift over time and need to be recalibrated periodically. A service engineer will know exactly how to do this. A less likely but possible cause is that your Bo field contains significant inhomogeneities due to improper adjustment of shims, gradient offsets, or presence of a ferromagnetic foreign object. Once again, a service engineer should be consulted who can identify and fix this problem. ADE |
|||