Diffusion MRI - Printable Version +- Unfortunately, the MRIQuestions.com forum became unmanageable due to spam attacks. I recommend an alternative resource: www.imagingQA.com (http://mriquestions.com/forum) +-- Forum: Pulse Sequences (Basic) (/forumdisplay.php?fid=5) +--- Forum: Diffusion imaging (/forumdisplay.php?fid=22) +--- Thread: Diffusion MRI (/showthread.php?tid=3440) |
Diffusion MRI - suren1111 - 04-15-2015 10:48 PM Dear Admin, I would like to know types of Diffusion MRI. I would also like to know Diffusion Kurtosis Imaging. ![]() RE: Diffusion MRI - aelster - 06-16-2015 05:49 PM Dear Suren111, Diffusion kurtosis imaging is still in the research stage, but very interesting. I didn't include it in the routine Q&A's but may consider putting it into an advanced section in the future. The basic idea is that the diffusion rate, even along a fixed direction, is not necessarily constant. Experimental evidence shows that water molecules tend to slow down faster than expected by a simple gaussian curve the further they are away from their origin. This can be expressed by a kurtosis factor, K, in that direction. I have attached a recent review article from the AJR you may find helpful. Allen Elster |